Re: Revisions to the high-Z impurity correction table
- To: Tom Banks <tbanks@socrates.berkeley.edu>
- Subject: Re: Revisions to the high-Z impurity correction table
- From: Claude Petitjean <claude.petitjean@psi.ch>
- Date: Fri, 18 Aug 2006 16:02:42 +0200
- Cc: Steven Clayton <smclayto@uiuc.edu>, Fred Gray <fegray@socrates.berkeley.edu>, Tim Gorringe <gorringe@pa.uky.edu>, David Hertzog <hertzog@uiuc.edu>, Peter Kammel <kammel@npl.uiuc.edu>, Malte Hildebrandt <malte.hildebrandt@psi.ch>, Rene Prieels <R.Prieels@fynu.ucl.ac.be>, Francoise Mulhauser <Francoise.Mulhauser@psi.ch>, Bernhard Lauss <lauss@berkeley.edu>, Brendan Kiburg <kiburg@npl.uiuc.edu>, Peter Winter <peter.winter@psi.ch>
- In-reply-to: <Pine.SOL.4.56.0608171134500.10980@socrates.Berkeley.EDU>
- Organization: psi
- References: <Pine.SOL.4.56.0510171101290.10438@socrates.Berkeley.EDU> <43C59BF1.70802@npl.uiuc.edu> <43C7FE61.3030209@npl.uiuc.edu> <Pine.SOL.4.56.0608171134500.10980@socrates.Berkeley.EDU>
- User-agent: Mozilla/5.0 (Windows; U; Windows NT 5.1; en-US; rv:1.6) Gecko/20040113
Dear Tom,
I studied today your memo on improved impurity corrections. - I think you
made indeed a comprehensive and quite perfect job!
Here are 2 considerations I made:
1) What would be the additional correction if the muon capture rate on the
proton would also be incuded in your formalism?
- basically, the free muon life time l_o needs to be replaced in your
parameter x by the real disapearance rate (l_o+L_s),
where L_s is our singlet capture rate, about 0.0015 of l_o.
- since x appears more or less linearly in the formula for Y, your
yield gets
just reduced this relative amount of 0.0015.
- this is indeed a negligible contribution.
2) You assume that in our run8 clean fill conditions we had mainly impurity
captures on Nitrogen. This was however a controversial issue, since
the Chemists analysed about 10 ppb Nitrogen - while the capture yield
suggested rather some 100 ppb.
In runs 9 and 10, using a humidity sensor, we concluded so far that at
least half, but very likelly 90% of the observed impurities come from
humidity (Oxygen).
I do not see a difference to run 8, since in all these runs the Nitrogen
was efficiently cleaned by the CHUPS system.
Therefore, would it not be more appropriate to take from table 3 for
delta_r/Y
the Oxygen value (0.800) rather than Nitrogen (0.823), or at least a mix?
Fortunately, the difference between these values is rather small.
Thats all comments that I have. I'm looking foreward to see in due time some
writeup of your work!
With best regards
Claude
Tom Banks wrote:
>Dear MuCappers,
>
>I have just posted a memo on my homepage which describes my revisions to
>Peter's longstanding table of numbers pertaining to the high-Z impurity
>correction:
>
> http://weak0.physics.berkeley.edu/weakint/research/muons/private/tbanks_dir/Documents/high-Z-table-revised.pdf
>
>I still have to finalize some literature references, but the body of the
>document is essentially complete and ready for inspection.
>
>As you will see, the changes to the values of the important quantity
>(delta r)/Y are not large, but they are large enough to affect the size of
>our high-Z correction for Run8 by roughly 1.5 Hz. I would greatly
>appreciate it if you could find the time to peruse the document and see if
>it all make sense. I think it is especially important that Steve and
>Peter K. inspect the calculations so that we can establish a consensus on
>the table's values.
>
>Regards,