[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]

EH-EH rates



Hi, Peter,

I have implemented a simple EH-EH trigger in the Berkeley analysis software.
I masked anode 48 (the "hot spot") but otherwise left things as naive
as possible, projecting onto the time axis (only) and looking for two EH
hits separated by at least one TDC400 tick.  The rates were then as follows:

Run    Condition               trigger rate [Hz]
                               muSC PP    muSC+muPC PP
23433  mu+ unkicked            220         88
22941  mu- unkicked            192         69
26434  mu+ unkicked             84         35
26769  mu- kicked              527        199
27733  natural H2 kicked       133         52

It is extremely important to have the best possible pileup protection 
for this analysis.  I think that you will have to include the muPC in
the trigger.

Once muPC1 has been included, the rate is dominated by events like the 
one I've attached, where there is a blue pixel along the green track.  
The probability of such events is strongly dependent on exactly how 
the E_high threshold is set relative to the gas gain, which explains the
large changes between runs.

However, we are not talking about an unworkable rate for the (mythical) 
new FADC hardware.  With appropriate tuning of the thresholds, we should be 
able to get 50-100 Hz, which should not be too much of a problem to read out 
and store, assuming some compression.

Thanks,

-- Fred

-- Fred Gray / Visiting Postdoctoral Researcher                 --
-- Department of Physics / University of California, Berkeley   --
-- fegray@berkeley.edu / phone 510-642-2438 / fax 510-642-9811  --

Attachment: cMuEventDisplay_1.ps
Description: PostScript document